Appropriate Handling of Out-of-Calibration Equipment
Calibration data are the key to preventing and handling out-of-calibration equipment.
Miriam Boudreaux
Mon, 10/05/2009 - 04:00
  • Add new comment
If you have ever found equipment that is out of calibration, then you know it is not something to take lightly. Whether you manufacture children’s toys or automobile tires, you know that the implications and ramifications of the decisions you make can be devastating for your company. Although the requirements from the ISO 9001 standard regarding equipment found to be out of calibration are simple and succinct, this is not something to take for granted. If you ensure that the processes for handling nonconforming equipment are in place and if you take into consideration the steps provided below, you will be ready to handle and perhaps avoid out-of-calibration conditions.
Equipment found out of tolerance
When calibrating your equipment and finding it to be out of tolerance, ISO 9001 requires you to consider the product that was inspected with such equipment as suspect product. Aside from quarantining the equipment for further adjustments and calibration, the first question you need to ask is: Does the calibration data suggest the equipment was broken, minimally out of tolerance, or grossly out of tolerance? Was it out of tolerance in the range in which it was used?
Submitted by davo@kapro.cn on Mon, 10/05/2009 - 18:15
you can also analyze the data on previous equipment checks
Submitted by Dave Gentile on Tue, 10/06/2009 - 07:35
Product Disposition
Miriam,
Your article is outstanding, but I disagree with some subtleties. I don't know what the current wording in ISO 9001 is, but as I recall the 1994 version said that it was necessary to analyze the degree and kind of calibration nonconformance, impact on past product acceptance, and whether product disposition was required. You are implying that a recall or reinpsection is mandatory:
PP "• If the product has already been shipped, the standard requires that you have a process in place for how you are going to recall it. If it was already delivered to the customer, the question is: How you are going to ask for the product back to inspect or test it again?"
PP I think the analysis has to be done carefully and ethically and the product disposition plan executed well - be it a full recall or no action at all. A company that goes beyond the binary mentality of inspection and truly controls its processes will probably not face this calibration scenario, but if they do they will likely have (a) processes that conform to standards regardless of inspection issues and (b) the knowledge required to (hopefully) arrive at the conclusion that the product is fine despite the calibration nonconformance. For instance, if one knew that a process was incapable of producing nonconformances (perhaps limited to the range of the equipment nonconformance), no action other than reporting would be necessary. Examples would be use of robust design practices like design for assembly, classification of characteristics, failsafing, consideration of the out-of-tolerance range on the gaging vs. the range at which the product inspection was done (forgive me if you already mentioned that).
PP Also, 10 years ago I found the QS-9000 MSA manual to be an excellent guide to running a strong measurement system, as it describes ways to assess gage capabilities that are overlooked by ISO 9001. I imagine the current version is even stronger.
Submitted by Miriam Boudreaux on Wed, 10/07/2009 - 20:56
In reply to Product Disposition by Dave Gentile
Product disposition
Hi Dave. Thanks for reading the article. You are correct in that recall or inspection is not mandatory. I did pose the question of assessing the confidence level in the equipment and subsequent results, but I can see the that the actions I suggested leaned towards the assumption that the monitoring or test activities using the equipment had an impact on product acceptance. Thanks for your in depth comments.
Submitted by jasonw on Fri, 05/31/2013 - 09:56
Broken equipment
I was wondering if you could share your take on broken or damaged equipment. For instance, you are 4 months into a 12 month cal cycle of an electrical meter and the inspector drops it and breaks it, maybe your 10 months in.  Regardless, for the sake of this discussion lets pretend that the inspector was dilligent and honest and submitted a written account of the occurance when it happened.  Maybe e-mailed quality or used an internal form, whatever the process is.  The bottom line is you have a clear record of when and how it occured.
In my view you should use the nonconformane process to document the removal of the asset from your system and the replacement if necessary but then you are left with a situation where you cannot validate previouse measurements through As-Found data at the closure of the calibration cycle.
I work for a calibration company so every measurment is clearly documented and I have the ability to look for trends in data suggesting it may have been producing false readings.  I can look at the history of the instrument and see if there are accounts of as found failing conditions or reason to believe that it would have been out of tolerance.  To me, that is the deviation from program, failing to compete as calibration cycle.  I've worked for a few different companies, and they all seem to have an approach to this matter however no one has ever really stated their method was based on any kind of official guidance beyond an internal procedure.
It doesn't seem realistic to take a guilty until proven innocent approach so to speak but on the other hand there should be some level of confidence the instrument was believed to be accurate during its usage.  In my case the last time the instrument was used it might have been thousands of miles away or in a different country even so I find myself searching for a cost effective approach to resolution without physical validation, or a process defending that physical validation was not deemed necessary.
I look forward to your reply,
Jason
Submitted by Miriam Boudreaux on Sat, 06/22/2013 - 19:02
In reply to Broken equipment by jasonw
Broken equipment
Hi Jason, Thanks for reading the article. I will attempt to answer the best way possible:
Actually what standard suggest is that you take action when the equipment "as received condition" during calibration was found out of tolerance. Reason being as you stated is to see how much product may have been inspected when the instrument was out-of-tolerance. So on you first sentence "you are 4 months into a 12 month cal cycle of an electrical meter and the inspector drops it and breaks it, maybe your 10 months in. Regardless, for the sake of this discussion lets pretend that the inspector was dilligent and honest and submitted a written account of the occurance when it happened. Maybe e-mailed quality or used an internal form, whatever the process is. The bottom line is you have a clear record of when and how it occurred". In my opinion, there was nothing to do except to change the status of your equipment on your Calibration log or register from "active" to "inactive" or "broken/retired".
You are correct in that nonconformance process should be invoke only when have suspicion that good product may have been inspected with out-of-tolerance equipment.  In your case, as you well stated "In my view you should use the nonconformane process to document the removal of the asset from your system and the replacement if necessary but then you are left with a situation where you cannot validate previouse measurements through As-Found data at the closure of the calibration cycle" that was not necessary, because there was never a doubt that product was inspected with this broken equipment.
Then you go on saying "I work for a calibration company so every measurment is clearly documented and I have the ability to look for trends in data suggesting it may have been producing false readings. I can look at the history of the instrument and see if there are accounts of as found failing conditions or reason to believe that it would have been out of tolerance. To me, that is the deviation from program, failing to compete as calibration cycle. I've worked for a few different companies, and they all seem to have an approach to this matter however no one has ever really stated their method was based on any kind of official guidance beyond an internal procedure."  Again, to me, there was no need to document, prove or trend anything.  the equipment broke, is now retired and future checks will be done using a new equipment.  But when the equipment was used to inspect product, it was indeed working fine.
Finally: "It doesn't seem realistic to take a guilty until proven innocent approach so to speak but on the other hand there should be some level of confidence the instrument was believed to be accurate during its usage. In my case the last time the instrument was used it might have been thousands of miles away or in a different country even so I find myself searching for a cost effective approach to resolution without physical validation, or a process defending that physical validation was not deemed necessary". I totally agree with you.  There is no need to prove anything. I will think any auditor asking you to document through a nonconformance that the equipment broke, is not valid, required nor value add. Only if you would have found the equipment broken, without knowledge of when it broke, then I would think that it may have some merit.  However if the equipment is broken to the point that it cannot be used for its primary functions- and there is no way that it would have been used for measuring- then, again, I would not document it status through a nonconformance.
Hope this helps, if not just email back.
Miriam
Submitted by Mark S on Wed, 09/23/2015 - 08:37
Out Of Calibration Is Not The Same As Out Of Tolerance
This article is a few years old, but people are still linking to it as it was e-mailed to me today.
Some good points are made, but the terminology can get slightly confusing.  Out of tolerance and out of calibration are two separate issues that don't necessarily overlap.  Out of calibration means that the M&TE was not recalled when it was due for calibration.  While there is some risk this equipment will be recovered and be out of tolerance, this is a small risk that is mostly independent of the equipment's due date.  Out of tolerance can occur any time the equipment is used but is more likely the result of damage, misuse, wear, or other mishap, assuming the equipment was stable to begin with.  To that end, the data absolutely helps in making that determination.  Data ultimately determines whether calibration frequencies are adequate to reach a desired reliability goal.  A popular target is 95% reliability.
Ultimately I'm not sure if out of tolerance conditions can be entirely prevented by data, but the risk can certainly be minimized.  A calibration system should be robust enough to handle this kind of "what if" even though it is unlikely to occur.
Submitted by Swati Malik (not verified) on Thu, 07/26/2018 - 00:19
Procedure for the disposition of affected product if calibration
Kindly suggest- Procedure for the disposition of affected product if calibration fails..?
Submitted by John Evans (not verified) on Tue, 08/06/2019 - 12:22
Out of Service equipment Instruments
Hi,
If an equipment/instrument is out of service, such as an instrument that produces DI or WFI, does that instrument need to have the DI or WFI sampled/tested if it is governed by an existing SOP that has a sampling schedule?
How about calibration or PM?
What FDA Reg. governs this situation?
Thanks
3 + 15 =